The next technique for refining the accuracy and precision of this method involves doing everything possible to fill the flask consistently with the same volume of water, both during calibration and when measuring the tank water density. There are two aspects to this. One is really just basic good measurement practices when reading the meniscus, but I will now elaborate on the specifics of how to do this. The other addresses a small but very real variability that can be minimized when filling the flask.
A volumetric flask is designed to allow a person to fill it with a known volume of liquid with very good precision. But for the precision required to perform the Specific Gravity calculation, extra care must be taken. I have previously linked to a NIST document describing the
Selected Procedures for Volumetric Calibrations. In this post, I'd like to draw the reader's attention to pages 19 and 20 of that PDF (GMP Page 1 of 4 and GMP Page 2 of 4 in the document). Those pages describe best practices for reading a meniscus. For this method, I suggest using "Option A" for how to read the meniscus. I have found that it is indespensably helpful to use a black/white reading card as described to get a very good view of the true bottom of the meniscus. I also want to emphasize attention to detail when it comes to setting the meniscus relative to the calibration line on the flask. In the NIST document's diagram labeled "FRONT VIEW", the calibration line is shown as though it is an infinitely thin line, but in fact, the calibration line on volumetric flasks has a definite thickness, of course. I have found that the thickness of a typical 100 mL Class A volumetric flask's calibration line corresponds to approximately 0.04 mL of volume, or about 1 drop of water. This amount of error in filling the flask corresponds to an error of about 0.5 PPT in the resulting salinity calculation, so in order to achieve the desired improvement in salinity resolution using this method, even that small amount of error should be avoided. While I don't in any way disagree with the NIST method described, I do wish to refine the description with my own images, intended to clarify how to properly set the meniscus.
Here is an image of a meniscus properly set using "Option A", with the darkness of the meniscus that would be seen when using the black/white card, and a red line indicating how the bottom of the meniscus should be halfway between the front and back parts of the calibration line when viewed from just beneath the calibration line, as described in the "Option A" part of the NIST document:
In this image, the higher, darker grey part of the calibration line is the front side, and the lower, lighter grey part is the rear side. Now, the clarification I'd like to add is that as you raise your eye relative to the calibration line, the bottom of the meniscus should remain halfway between the lower portion of the front side of the line, and the upper portion of the rear side of the line, as shown in the next two images:
If you have the flask filled just a little too full, then the meniscus will tend to disappear behind the front side of the line before you get to the orientation in the last image, and if you have the flask just a little under-filled, then it will still be clearly visible at that angle, and it may even start to obscure the rear side of the line. If you have filled the flask just right, then when you get to the right angle, the bottom of the front side of the line, the bottom of the meniscus, and the top of the rear side of the line will all line up exactly. For my older eyes, I find a moderately strong hand lens helps immensely to see this well. And again, don't forget to use the black/white card to clearly show where the bottom of the meniscus really is. You might think you can see it clearly without the card, but in practice, the card really does help to see it properly under variable lighting conditions.
It is important to use a syringe or pipette of some kind (a Salifert test it syringe with the tip on works nicely) to add or remove the tiniest amount of water necessary to set the meniscus in just the right place to achieve what I have shown above for consistent measurement.
There is one other minor but very important detail to pay attention to in order to get consistent measurements between different flask fillings, and that is the drops that may be left adhering to the neck of the flask above the calibration line. In the NIST document, there is a section that describes the proper cleaning of glassware to ensure that all the liquid flows off of the surfaces of the glass without leaving drops behind. This cleaning process involves using a hot solution of sodium dichromate in fuming sulfuric acid to do this cleaning! I'm not into having such dangerous solutions in my home, and I doubt any hobbyists reading this are, either. Even with such extreme cleaning methods, the NIST document does also describe one of the uncertainties in their calibration method being the unknown amount of liquid that remains clinging to the neck of the flask. In practice, no matter how scrupulously I try to clean my glassware (I do use reasonably strong sulfuric acid and sodium hydroxide in the effort), I still get some drops clinging the the neck of the flask when I fill it. I estimate the extra volume of these drops can be as much as 0.03 mL or so. So, my solution, which I have found works very well, is to simply wipe the inside of the neck of the flask, after I've filled it and set the meniscus, with a rolled-up paper towel. I fold the paper towel in half, and then roll it up so that the folded edge is at one end of the roll, and insert it into the flask folded edge down, so that there are no fibers sticking down that might touch the meniscus, and wipe the inside of the flask as close as I dare come to the meniscus without touching it, rotating it as I insert it. This works wonders to reduce the uncertainty of extra water clinging to the neck.