I'm thinking about getting one of the Hanna Marine Alkalinity checkers to replace the Salifert titration test. Hanna makes two of these - one's the HI772, and one's the HI755. I'm interpreting that the only difference between these two colorimeters is that the HI772 reads in dKH, and the HI755 reads in ppm, but otherwise the principle of the test remains the same.
My purpose for doing this is to increase the precision of my alkalinity determinations. The Salifert test yields a precision of (at best) +/- 0.3 dKH, and in actual use, the precision's probably no better than +/- 0.6 dKH because the actual endpoint is somewhat open to interpretation.
However, I'm aware that precision isn't the same as accuracy, so I'd like to get analyzed standards with the Hanna. It appears that Hanna's HI755-11 standard set is intended for the HI755, while they sell another standard set (the HI772-11) for the HI772 meter. Does anyone know why there would be two separate standard sets? I'm interpreting (perhaps incorrectly) that the wavelength for the two colorimeters is identical, so it's puzzling why they would supply two different calibration sets.
I should note that I have the equipment, reagents and knowledge to do my own highly accurate alkalinity titrations using a standardized HCl solution, buret, pH meter and stir plate. So one solution to the accuracy question would be to simply compare the Hanna reading to a titration of the same sample. But that's a bit inconvenient, so if anyone has thoughts about these Hanna calibration sets, I'm all ears.
My purpose for doing this is to increase the precision of my alkalinity determinations. The Salifert test yields a precision of (at best) +/- 0.3 dKH, and in actual use, the precision's probably no better than +/- 0.6 dKH because the actual endpoint is somewhat open to interpretation.
However, I'm aware that precision isn't the same as accuracy, so I'd like to get analyzed standards with the Hanna. It appears that Hanna's HI755-11 standard set is intended for the HI755, while they sell another standard set (the HI772-11) for the HI772 meter. Does anyone know why there would be two separate standard sets? I'm interpreting (perhaps incorrectly) that the wavelength for the two colorimeters is identical, so it's puzzling why they would supply two different calibration sets.
I should note that I have the equipment, reagents and knowledge to do my own highly accurate alkalinity titrations using a standardized HCl solution, buret, pH meter and stir plate. So one solution to the accuracy question would be to simply compare the Hanna reading to a titration of the same sample. But that's a bit inconvenient, so if anyone has thoughts about these Hanna calibration sets, I'm all ears.

The difference is in the units reported as the op noted.