I installed the Apex Gold Package in Jan 2015. So far things have been working well except for my problems in connecting my radions to the WXM module.
Since last 2 weeks my conductivity reading started going up from 35 to 36 then 37 and ended up at 38 - when I decided to investigate.
1. Checked the reading of my tank with a calibrated refractometer - reading was 35.
2. Dried the probe for re-calibration for 3 hours.
3. Followed the re-calibration process and the reading in the re-calibration solution was 34.9.
4. Add the probe back in my sump the reading drops to 31.
5. Put the probe back in the test solution - reads 34.6.
Waited for a day. Followed the same above steps after drying the probe for longer period of 6 hours. Got exactly same results. My temp compensation is 0.0 and I have carefully cleaned the probe of any debris.
Waited for another 2 days. This time dried the probe for 2 days complete and followed the re-calibration process - exactly same results.
I re-calibrated my refractometer using the same solution I am using to re-calibrate the probe and it reads 35. I have run out of options to fix or troubleshoot this problem. Any help will be greatly appreciated !!
Since last 2 weeks my conductivity reading started going up from 35 to 36 then 37 and ended up at 38 - when I decided to investigate.
1. Checked the reading of my tank with a calibrated refractometer - reading was 35.
2. Dried the probe for re-calibration for 3 hours.
3. Followed the re-calibration process and the reading in the re-calibration solution was 34.9.
4. Add the probe back in my sump the reading drops to 31.
5. Put the probe back in the test solution - reads 34.6.
Waited for a day. Followed the same above steps after drying the probe for longer period of 6 hours. Got exactly same results. My temp compensation is 0.0 and I have carefully cleaned the probe of any debris.
Waited for another 2 days. This time dried the probe for 2 days complete and followed the re-calibration process - exactly same results.
I re-calibrated my refractometer using the same solution I am using to re-calibrate the probe and it reads 35. I have run out of options to fix or troubleshoot this problem. Any help will be greatly appreciated !!
.