Hi all, I received a good suggestion that I should post the fix that finally worked to get my Apex Salinity probe calibrated and working properly so here it is:
I followed the manual calibration, not the automatic calibration, instructions to the letter (copied below) as provided by Neptune. The difference this time was how well I cleaned and dried the probe prior to calibration. This included a good but gentle rinse in RO/DI water for about 30 seconds and making sure that the probe was wiped dry and had no lingering water inside the tube. A few gentle taps on a dry cloth helped to dislodge any remaining water. I then waited several minutes before attempting to calibrate, maybe 5-6.
From there I followed the prescribed steps and it all worked as described. Nothing magic here unfortunately but thankful that it now works and can be verified against other calibrations for accuracy notably my Hanna calibration tool and my local shop...
I hope this helps. Happy trails...
MANUAL CALIBRATION (credit to Neptune Apex Support team)
Step 1:
Start by logging into your APEX Dashboard and clicking on the
Three Gears Icon and then the
INPUTS Icon
Step 2:
Select the
SALT(BASE_COND) Option.
Step 3:
Ensure you have the Conductivity Probe securely plugged into the APEX.
Step 4:
Click on the
ADVANCED Tab option.
Step 5:
Scroll down and click on the
MANUAL PROBE CALIBRATION option.
Step 6:
Follow the steps it prompts you to do. Press the Calibrate button.
Air dry the Conductivity Probe
. (This can take 1 minute or less to dry) *note:Some customers have used canned air
Step 7:
Next Select Ok once the probe has settled
(the values not changing after 15-30 seconds)
*Note the typical settling number range is 60-90 when calibrating Dry
Step 8:
Select the Calibration Solution you are about to use.
*Note the Apex base unit can ONLY calibrate using 53,000, a PM2 module can offer other ranges
Step 9:
Dip the Conductivity Probe into that calibration solution. Ensure to gently swish the probe, so as to avoid micro bubbles on the surface of the probe.
This settling number may not completely stop. But, generally when it's settling down go ahead and click next.
(once the values hold from changing for 15-30 seconds)
*Note the typical settling number range is 200-900 when calibrating in the 53,000 Solution
Step 10:
Now you will get this message that you have successfully calibrated the probe
Step 11:
Once the calibration is complete your Salt reading on the APEX Dashboard should read around
34.7~35.3 inside of that packet of calibrating solution. In dry air, the probe will read
0. In the solution it will read
34.7~35.3. If it does then the probe is calibrated successfully. You are now complete!
Note: You can save the calibration fluid for a couple days if you seal up the solution packs when not in use.
Please send us the results of:
1) Settling number from step #7 - Dry
2) Settling number from step #9 - In the Cal solution
3) Settling number from step #11 - In the Cal solution
4) Settling number in the tank