The refractometer and reference solution is from BSR calibrated using the calibration solution which is 35 PPT at 53.0ms.
The salinity probe is the apex Neptune that came with the system and was calibrated twice the last time with a apex technician using their reference solution.
First thing I did was calibrate the refractometer. Then I cleaned it with ro water and then put a sample of a tank water. My numbers are off by more than 2 points.
I checked for micro bubbles on the salinity probe by tilting it and shaking with no change.
I just need to establish a reference point of 35. If I use the apex as the reference and it’s wrong my swing could be 36 to 38 PPT if it’s correct no problem. If I use the refractometer it’s off and the apex is correct on my swing can be 31 to 33. So what I’m trying to say is my numbers would be between 31 & 38 which 31 is too low and 38 is too high. I also know consistency is the key and I don’t mind being off but I think this is just a bit too much.
Just had my lights come on.
As a test tonight I’m going to open up another bag of reference solution from Apex stick the probe in there and see if I get 35 I am also going to take some droplets from the same reference and see if I get 35 on the refractometer. If I get the same results then I know that I am getting some interference on the probe when it’s in the water. If I don’t at least I will know which device is off.
Tony T np it happens
The salinity probe is the apex Neptune that came with the system and was calibrated twice the last time with a apex technician using their reference solution.
First thing I did was calibrate the refractometer. Then I cleaned it with ro water and then put a sample of a tank water. My numbers are off by more than 2 points.
I checked for micro bubbles on the salinity probe by tilting it and shaking with no change.
I just need to establish a reference point of 35. If I use the apex as the reference and it’s wrong my swing could be 36 to 38 PPT if it’s correct no problem. If I use the refractometer it’s off and the apex is correct on my swing can be 31 to 33. So what I’m trying to say is my numbers would be between 31 & 38 which 31 is too low and 38 is too high. I also know consistency is the key and I don’t mind being off but I think this is just a bit too much.
Just had my lights come on.
As a test tonight I’m going to open up another bag of reference solution from Apex stick the probe in there and see if I get 35 I am also going to take some droplets from the same reference and see if I get 35 on the refractometer. If I get the same results then I know that I am getting some interference on the probe when it’s in the water. If I don’t at least I will know which device is off.
Tony T np it happens