Was able to do the calibration. It did measure the 40ppt and 8ppt stock solutions after calibration correctly so seemed that calibration was ok, but when I measured a reference fluid of 33ppt, it still showed 21.2 ppt, 2.12% and 1.014 for that 33ppt saltwater solution. So I assume it does not measure seawater ppt but some other salinity?
Found a "data sheet" of the product: http://www.tools.in.th/manual/Catalog-AR8012.pdf
What standard do you have that is 33 ppt? Maybe it was not made correctly.
You probably also calibrated the refractometer with it, so matching the refractometer may be circular logic if that standard is off.