I made up some DIY solution with the following from RHF's article (using expensive, presumed accurate scales, and 0 TDS water):
This 3.65 weight percent sodium chloride solution can be made by dissolving 3.65 grams of sodium chloride in 96.35 grams (mL) of purified freshwater.
I checked this solution in my refrac, it measured at about 30 PPT. I adjusted the refrac to read 35 PPT with this sample, and triple checked with further readings.
Next I checked my TLF Accurasea, this now (after the adjustment above) reads at 39 PPT. The sticker udner the bottle says 35.1 PPT.
So my question is which to believe?!
The background and reason I am checking is that i've always trusted the Accurasea solution, but it didn't match the result I got from my ATI ICP, and water from my LFS read at 1.020 when they told me it was 1.025 (IIRC). Next I got some sea water from the coast and checked that, it read at 1.020...
Is there another way to verify this reading?
edit: another interesting point, when calibrated with the DIY solution, RODI measures at zero, when calibrated with the Accurasea it is significantly off zero.
This 3.65 weight percent sodium chloride solution can be made by dissolving 3.65 grams of sodium chloride in 96.35 grams (mL) of purified freshwater.
I checked this solution in my refrac, it measured at about 30 PPT. I adjusted the refrac to read 35 PPT with this sample, and triple checked with further readings.
Next I checked my TLF Accurasea, this now (after the adjustment above) reads at 39 PPT. The sticker udner the bottle says 35.1 PPT.
So my question is which to believe?!
The background and reason I am checking is that i've always trusted the Accurasea solution, but it didn't match the result I got from my ATI ICP, and water from my LFS read at 1.020 when they told me it was 1.025 (IIRC). Next I got some sea water from the coast and checked that, it read at 1.020...
Is there another way to verify this reading?
edit: another interesting point, when calibrated with the DIY solution, RODI measures at zero, when calibrated with the Accurasea it is significantly off zero.