Hi R2R,
Yesterday I installed a new UV Sterilizer to combat some dinos. I noted the time I installed it so I could track results as 3:30pm (December 29th). Since that time I've noticed very erratic PH readings from my Apex PH Probe. Curious if anyone knows why this might be happening? I do have some calibration solution and plan on doing a cleaning and re-calibration in the next few days. Any other thoughts what could cause this? Below is a graph including the prior day and the time since installation. Notice the more frequent jumps up and down since 3:30 install.

Screen Shot 2021-12-30 at 11.40.32 AM.png
 
Move the Sensors away from the UV and UV Transformer. Also some LED Lights, DC Drive Pumps, can Cause bad readings from pH Monitors.
 
Even power cords crossing over the probe wires can cause problems, worth checking.
 
OK. That's a good tip. I might have moved things around too much in the equipment cabinet. I just untangled the PH probe a bit and moved the larger power items farther away. I'll check the graphs again in a few hours.
 