hanna checker question?

hotnickz07

Active Member
View Badges
Joined
Oct 29, 2013
Messages
368
Reaction score
179
Location
FL
Rating - 0%
0   0   0
Hey guys im having a hard time determining the correct reading on my hanna phosphate reader. If i quickly tap the button i get a reading of .09 bit i dont get a countdown. Just blinks a few times and get my reading. But if i longpress the button and get the 3 minute countdown going , im getting a 0.00. Whats going on here. What im i doing wrong??
 
actually, for Alk you don't hold it down but for phos you do hold it down. The phos takes the full 3 minutes to evaluate the sample and give you the correct value.
 
copy paste on what I do and applies to the 713 and 736

736 manual
http://shop.hannainst.com/downloads/...6_13_05_10.pdf
713 manual
http://shop.hannainst.com/downloads/...t713_03_13.pdf
I always make sure the do the same steps every single time and have always gotten consistent results. Others have had bad reagents in the past but I haven't had the problem yet but it's always a possibility with any test kit.




I have a pretty good routine down with this checker and do the same with the 713 checker as well. Always handing with a lint free cloth and never touch with fingers though all steps.




1. rinse out the vial with tank water several times and inspect vial for any blemishes.

2. Fill that vial with 10ml of tank water taking care not to have any bubbles on the sides (if filtered can skip step 4.)

3. Get a packet out and flick it a few times to loosen up the powder and move away from the area I'm going to cut in a curved pattern as indicated on the packet (I was doing this before they put that indicator on there as well)

4. Set the vial down with tank water and wait at least minutes doing other stuff ( I used filter out the sample with a coffee filter but this seems just as good and I'm usually doing several things at the same time)

5. Turn the checker on and wait for it to ask for C1

6. Place the vial in with the 10ml mark facing me

7. Cut open the reagent packet and pinch it open so the bottom part is creased to help funnel the powder out into the vial

8. Press the button to read C1

9. As soon as it asks for C2 I pull the vial out (never touching with my fingers at any stage of this using a lint free cloth) and add the reagent

10. Close the vial up and look at the time and gently start mixing the reagent into the water. Taking care not to form bubbles on the insides of the glass. Alternating between slowly inverting and gently swirling

11. As soon as the time changes on the 2 minute mark I press and hold the button to get the 3 minute timer.

12. Continue to gently mix the reagent for 30 more seconds inspecting for any bubbles on the sides or accidental blemishes

13. Place the vial back into the checker and wait for the count down.

14. Rinse immediately after getting the reading.




I find that the reagent for the ULR 736 checker dissolves a lot better then the 713 Low Range checker. Also the end result value is displayed a lot longer on my 736 ULR checker.
 
Cool beans!!. Thanks just so confused as why im getting 2 readings. So it seams its the press and hold method to get the countdown going and not the other one
 
Dang jason!! You are more ocd than me.. and thats hard. Lol
 

IF YOU HAD TO TAKE A REEFING EXAM, WOULD YOU PASS?

  • Yes!

    Votes: 32 45.7%
  • Not yet, but I have one that I want to buy in mind!

    Votes: 9 12.9%
  • No.

    Votes: 26 37.1%
  • Other (please explain).

    Votes: 3 4.3%

New Posts

Back
Top