I can't seem to get a consistent reading from my Hanna LR Cu test. Does anyone else have this problem?
This morning I asked about QT problems and someone suggested copper. So I pulled out my handy Hanna low-range Copper checker and got results from 0 to 178 in a series of 6 tests.
Just now I tested my RO/DI, salt mix and DT with similar widely varying results (table below).
Procedure:
I was very careful to do all the tests the same except that they used different vials. So I re-washed 2 vials re-tested my RO/DI, two tests each, thoroughly washing the vials between tests and keeping track of which vial went with each test. The first vial got two 0 results -- good, I just hadn't washed my vials well enough. The third test (now with the other vial) came up 0 -- good, all going as expected. I re-washed that vial and did a second test with it -- 100ppb.
I give up. If I can't get consistent results when being as careful as I can possibly be, what good is this overpriced piece of rubbish?!?
Does anyone else have this problem? I see multiple reports that the HR Cu tester is unreliable at the low end. Is the same true for the LR checker, and might that explain my results?
I also double-checked and all my reagents are LR, not expired and from the same batch:
I'm not out of Cu LR reagent. I almost ordered more to keep testing, but I'm not sure I have faith in Hanna anymore and I'm reluctant to give them more money.
Here's my data:
This morning I asked about QT problems and someone suggested copper. So I pulled out my handy Hanna low-range Copper checker and got results from 0 to 178 in a series of 6 tests.
Just now I tested my RO/DI, salt mix and DT with similar widely varying results (table below).
Procedure:
- Filled and emptied 6 vials with tap 10 times each. Dried with paper towel.
- For each test:
- Filled and emptied syringe used for filling vial with test water 3 times.
- Filled and emptied vial with test water 10 times.
- Rinsed vial cap with 1 full syringe (20ml) of test water.
- Filled vial using syringe and capped.
- Dried outside vial with towel and wiped down with lint-free cloth.
- Put vial in tester and pressed button for C-1 (per instructions).
- Carefully added reagent after cutting along dotted line and folding as shown in instructions.
- Mixed reagent for 15 seconds per instructions. I do shake for about 10 seconds because at low readings that's the only way to get all the reagent to dissolve. But I then swirl for 5 and tap the vial to release bubbles.
- Wiped down vial with lent-free cloth again and put into tester.
- Wait 45 seconds for bubbles to settle per instructions.
- Close tester and press button for C-2 (per instructions).
- Read results
I was very careful to do all the tests the same except that they used different vials. So I re-washed 2 vials re-tested my RO/DI, two tests each, thoroughly washing the vials between tests and keeping track of which vial went with each test. The first vial got two 0 results -- good, I just hadn't washed my vials well enough. The third test (now with the other vial) came up 0 -- good, all going as expected. I re-washed that vial and did a second test with it -- 100ppb.
I give up. If I can't get consistent results when being as careful as I can possibly be, what good is this overpriced piece of rubbish?!?
Does anyone else have this problem? I see multiple reports that the HR Cu tester is unreliable at the low end. Is the same true for the LR checker, and might that explain my results?
I also double-checked and all my reagents are LR, not expired and from the same batch:
I'm not out of Cu LR reagent. I almost ordered more to keep testing, but I'm not sure I have faith in Hanna anymore and I'm reluctant to give them more money.
Here's my data:
| Source | Cu (ppb) | Vial |
| RO/DI | 0 | A or B |
| RO/DI | 241 | A or B |
| Salt mix | 71 | C |
| Salt mix | 28 | D |
| DT | 27 | E |
| DT | 0 | F |
| RO/DI | 0 | A |
| RO/DI | 0 | A |
| RO/DI | 0 | B |
| RO/DI | 100 | B |

Not really practical or affordable for the change, test, iterate I need to do to verify if Cu is the problem in my QT, but I'll get my DT tested. That should hopefully rule out RO/DI failure and bad salt mix as sources if copper it is.