@living_tribunal
The meter gives reliable results if
1. use a syringe to measure 10ml into the test cuvette
2. Wipe the outside clean with a paper towel or cry cloth to remove smudges from the glass.
3. Start the meter, add the clean sample cuvette. I align the 10ml mark with the front of the meter.
4. When it asks for the second sample, remove the cuvette and add all of the envelope contents. Start a stop watch and shake for the full 2 minutes.
5. Put the shaken sample back into the meter being careful to align the 10ml mark again with the front of the meter.
6. Hold the test button down until it starts the 3 minute timer.
The test reagent colors are at there peak for correct readings at 5 minutes in the cuvette, so the two minutes of shaking + 3 minutes of timer give the most accurate results.
The repeated alignment of the cuvette helps prevent errors from light deflection as the test is performed.
The old #4 = Rinse the cuvette with Deionized water after the test is finished. I rinse my cuvettes and then fill them with DI before I store them. It helps prevent mineral build-up on the glass which can also impact test results.
The above information was gleaned from others on here who have posted about Hanna Checkers and getting correct results. IHTH.