Same question. Anyway, if you want to see how precise it is, have two vials ready. One just with tank water for your "blank". Use other for the "sample" (tank water and the reagent). Mix the latter thoroughly for a couple minutes, wipe it down thoroughly with a microfiber cloth or a tissue damp with RODI water then dry it, make sure there aren't air bubbles inside (if so, gently swirl), then let it sit for 3 minutes. Now do multiple tests. Power up the tester. Put the blank vial (also cleaned carefully) in when it says C1. Make sure the tester lid is closed tightly. Press the button to zero the tester. When it says C2, remove the blank and replace with the sample. Press the button once. Don't hold it which will set the timer for 3 minutes---the sample has already sat for three minutes. This will give you a reading. Cycle the power on the tester and repeat the blank zeroing and sample measurement again. Then do it a third time. You should have very similar if not exactly the same results from all three tests. If you're doing the three tests in quick succession, the longer time the sample vial has sat between measurements shouldn't matter much (as long as you shook the vial for 2 minutes and then let sit for 3 minutes at the beginning). The above method isn't meant to replace the standard testing protocol. Instead, it's meant to gauge the precision of the checker as it eliminates most of the variability that could be attributed to user error.
[BTW, in my experience, you don't need to align the test vial the same way in the checker every time. Also, once you clean the vials, try to only touch them by the cap to avoid smudges on the glass.]