In case you missed it - this is what Hanna technical support said (posted days ago)
I just called Hanna techincal support here are the answers:
1. The test is to be read 5 minutes +- 15 seconds (after the reagent is added to the vial.)
2. The vial needs to be mixed for "at least" 1 minute 45 seconds - preferably 2 minutes. (He said the 1:45 is for those who find it difficult to accomplish everything in the time allowed.). He also said they realize the shut off time is too short

3. Repeated testing - and tests done after longer than 5 minutes are not valid. The test is designed to be read at about 5 minutes.
4. If you want to do repeated testing, you have to obtain a new sample and repeat the entire test.
5. As mentioned earlier - its best to use the same cuvette for both the control and the test.