Hanna has to say this because they sell to everyone – people with functional RODI filters who maintain their units AND people who never change their filters and have "RODI water". If you have 0 TDS RODI water it should be fine....I never used distilled and (once I practiced) had no problem matching my test results to Salifert.
Distilled water is less of a question mark in terms of quality variability. Since almost nobody makes their own distilled that means it's less likely to be screwed up – it's all "professionally made".
One important thing to note is TDS is actually the least refine measurement for water purity.
A TDS meter is used to determine the amount of dissolved particles in the water, usually in parts-per-million or ppm. However this is not necessarily true. TDS meters are actually measuring EC or electrical conductivity.
EC is the measurement of how well a medium (usually water) conducts electricity. In general conductivity increases as the concentration of ionic dissolved solids increases. Using EC as a measure of TDS is non-specific for ion type. An ion is a charged particle present in the solution that contributes to the current flow. Conductivity measurements are used for additional units such as TDS and salinity. EC is commonly expressed as micro Siemens/per cenitmeter (μS/cm) or milli-Siemens/per centimeter (mS/cm). Units may be converted as:
1000 μS/cm = 1.0 mS/cm.
Most EC/TDS meters use percent (%) of the full scale (f.s.) to determine accuracy. This becomes very important in relation to low conductivity water where values as close to zero are preferred. Range is used to describe the minimum or maximum detection limits for that particular meter. Range is thus used in determining the accuracy of TDS meters. For example, a common accuracy statement for many EC/TDS meters is ± 2% f.s., with a range of 999 ppm TDS. In this scenario, the measure of uncertainty is 2 percent of 999 ppm.
0.02 X 999 ppm = 19.98 ppm
This would means your results could be ±20 ppm different then the results displayed. So if your meter was reading 30ppm the results could actually be from 10 ppm to 50 ppm.
Conductivity/TDS meters come in a variety of price points ranging in quality, accuracy and applications. If you are using a TDS meter for RODI water to be used in a reef tank, then you want a meter designed for low range conductivity. A goal of most reef keepers is to have source water as close to 0 ppm total dissolved solids as possible. Many hobbyists use their TDS meter to determine when it’s time to change their RODI cartridges. If your TDS meter has a range of 1000 ppm or more, then you can expect your results to never be more accurate than ±20 ppm, assuming a 2% f.s. accuracy statement.
For spot-checking your system we recommend the
HI98308 Pure Water Tester. This handheld meter is temperature compensated, has a maximum detection limit of 99.9 µS/cm (approx. 50ppm TDS), accuracy of ±2% f.s., and can be manually calibrated with
84 µS/cm buffer. This meter can be used for checking your reservoir or by taking a sample directly from your RODI output. When using the Pure Water Tester, a reading of 20.0 µS/cm would equate to 10ppm TDS, and your measure of uncertainty would be from 19.6 µS/cm to 20.4 µS/cm.
I bring this up because the HI758 Calcium Checker does need
pure deionized water in order to function within its accuracy statement.