I have a TropicMarin hydrometer, about 12" long, and measures to 4 decimal places. Supposed to be lab grade, but I use a refractometer.
One thing we learned recently is that calibrating with distilled/RODI water can skew the actual SG by several thousandths. I recently started using Pinpoint 1.0264 (35 ppt) calibration solution and found my SG was at 1.021-22 instead of the 1.026 I thought it was. After Pinpoint calibration I put some distilled water on the refractometer and it actually read below zero.
Dave
One thing we learned recently is that calibrating with distilled/RODI water can skew the actual SG by several thousandths. I recently started using Pinpoint 1.0264 (35 ppt) calibration solution and found my SG was at 1.021-22 instead of the 1.026 I thought it was. After Pinpoint calibration I put some distilled water on the refractometer and it actually read below zero.