Okay guys, so the results are in and the Milwaukee is actually spot on. I got another refractometer standard solution today for 35 ppt to use in the testing as well. I tested everything multiple times on the Milwaukee VS manual refractometer and it turns out the standard fluids are off. Measurements for all fluids and tank water were identical on the Milwaukee VS manual refractometer.
The calibration fluid measured 0 ppt and 1.000 sg on both refractometers.
My rodi water measured at 0 ppt and 1.000 sg on both refractometers.
The 1.025 sg standard solution measured at 1.023 sg on both refractometers.
The 35 ppt standard solution measured at 37 ppt on both refractometers.
My tank water measured 35 ppt on both refractometers.
Conclusion: Make your own standards.