New Milwaukee MA887 Digital Seawater Refractometer - results question

  • Thread starter Thread starter Edp251
  • Start date Start date
  • Tagged users None

Edp251

Active Member
View Badges
Joined
May 31, 2018
Messages
257
Reaction score
901
Location
Hingham
What state or country do you live in
Massachusetts
Rating - 0%
0   0   0
Hey All,

Received my new salinity checker and tried it out this morning and got an alarming value of 1.03...

To be clear, tank and corals all look great so not an emergency but I am curious about which methods to trust.

Over the last few years I have used the same refractometer calibrated before each use. Since set up in December, my current tank has consistently read 1.026 - 1.0265 leaning a bit towards the high side.

When I got the the 1.03 reading on the Milwaukee, I started testing things:

1. Both my refractometer and the Milwaukee read SG 1.00 using the same steam distilled water standard
2. With my bottle of 1.025 standard - the refractometer read 1.025 but the Milwaukee read 1.028
3. With the Milwaukee standard - the refractometer read 1.02 and the Milwaukee read 1.025

So, my assumption is then that my bottle of 1.025 standard is actually closer to 1.028 due to age, precipitation, evaporation etc.. and this could be confirmed on someone else's refractometer or Milwaukee.

This means my tank salinity has crept up and needs to be carefully lowered.

Would you assume the same?
 
Had a bad batch of standard with my Milwaukee Refractometer. They sent me another. All is good
 
Interesting! I've got a digital Milwaukee refractometer. I don't have another device to compare measurements. Thought about grabbing a cheaper refractometer to check.
 
Hi Snuggs,

What did your 'bad' batch of standard read on the Milwaukee?

Thanks!
 
You can make your own standard:

Refractometers And Salinity Measurement
http://www.reefedition.com/refractometers-salinity-measurement/



Do-it-yourself Refractometer Standards

In a previous article I described how to make a do-it-yourself refractometer standard matching 35 ppt seawater, and I will just summarize that recipe here.

To provide a standard for refractometers requires a solution whose refractive index is similar to normal seawater. Seawater with a salinity of 35 ppt has a refractive index of 1.3394. Likewise, the refractive index of different sodium chloride solutions can be found in the scientific literature. My CRC Handbook of Chemistry and Physics (57th Edition, Page D-252) has such a table. That table has entries for 3.6 and 3.7 weight percent solutions of sodium chloride that span the value for normal seawater. Interpolating between these data points suggests that a solution of 3.65 weight percent sodium chloride has the same refractive index as 35 ppt seawater, and therefore can be used as an appropriate standard (Table 5).


This 3.65 weight percent sodium chloride solution can be made by dissolving 3.65 grams of sodium chloride in 96.35 grams (mL) of purified freshwater. This recipe can be scaled to any appropriate size if suitable instruments are available (36.5 grams in 963.5 grams (mL) of water, 0.365 grams in 9.635 g (mL) of water, etc.).

This concentration roughly corresponds to ¼ cup (73.1 g) of Morton’s Iodized Salt dissolved into two liters (2000 g) of water (giving very slightly more than 2 L of total volume).

For a rougher measurement in the absence of an accurate water volume or weight measurement:

  1. Measure ¼ cup of Morton’s Iodized Salt (about 73.1 g).
  2. Add one teaspoon of salt (making about 79.3 g total salt).
  3. Measure the full volume of a plastic 2 L Coke or Diet Coke bottle filled with purified freshwater (about 2104.4 g).
  4. Dissolve the total salt (79.3 g) in the total water volume (2104 g) to make an approximately 3.65 weight percent solution of NaCl. The volume of this solution will be slightly larger than the Coke bottle, so dissolve it in another container.
[Note: the standard described here using soft drink bottles is subject to variation in the volume of the bottle. It turns out that such bottles can vary in total volume, and this can lead to at least a one ppt error in the salinity of standards matched to seawater of 35 ppt salinity. Standards made with accurate measurements of salt and water, however, will accurately match 35 ppt.]
 
I have found that my milwaukee consistently reads .002 high, even after regular calibration and being sent back for warranty repair. Zeros fine, but 1.026 always reads 1.028. I just aim for that with it now. Been fine that way.
 