I'm not sure if this is helpful, because I didn't take the comparable spectral measurements, but it's what I have, so I wanted to share. Take it for what it is.
Using the Reefi Uno (1st gen), I measured individual channels with the PARwise and compared it to the review Dana Riddle took using the Ocean Optics USB2000 spectrometer (attached).
400nm
Dana - 398
PARwise -385
420nm
Dana - 420
PARwise - 425 - 430 (bounced between the two)
430nm
Dana - 430
PARwise - 435
450nm
Dana - 450
PARwise - 444
470nm
Dana - 472
PARwise - 472 - 477 (bounced between the two)
All On
Dana - 420 peak
PARwise - 444 - 446 peak (bounced between the two)
I did not measure lime, amber, warm white, or cool white.
One peculiar trait I noticed, when I turned on my Aquatic Life Coral Cover fixtures I use for fill light, which has a peak at 620nm, the PARwise would detect the peak only if my 430nm channel was turned off (nothing else changed). As soon as I turned the 430nm channel on, the 620nm peak flattened. I checked the other channels, but it only happened with the 430nm.