When testing My salinity the first reading will usually be high , followed by the reading dropping slightly every time . Has anyone else had this issue or is my refractometer ready for the bin ? Have calibrated it to 0 using R.O but it always does this. First test can be as high as 1.032 and generally gets to the stage where it will settle between 1.025 - 1.027 for half a dozen tests in a row but just makes me uneasy that I'm getting different readings. But always starting high and gets gradually lower ?
 
When testing My salinity the first reading will usually be high , followed by the reading dropping slightly every time . Has anyone else had this issue or is my refractometer ready for the bin ? Have calibrated it to 0 using R.O but it always does this. First test can be as high as 1.032 and generally gets to the stage where it will settle between 1.025 - 1.027 for half a dozen tests in a row but just makes me uneasy that I'm getting different readings. But always starting high and gets gradually lower ?
Calibrating with RODI may not be the best way to go. As far as I understand, it is recommended that you purchase a calibration solution at 35 ppt and use that instead of RODI.
 
Calibrating with RODI may not be the best way to go. As far as I understand, it is recommended that you purchase a calibration solution at 35 ppt and use that instead of RODI.
I understand that but even still it doesn't explain why my test results are slowly dropping from 1 minute to the next ?
 
When testing My salinity the first reading will usually be high , followed by the reading dropping slightly every time . Has anyone else had this issue or is my refractometer ready for the bin ? Have calibrated it to 0 using R.O but it always does this. First test can be as high as 1.032 and generally gets to the stage where it will settle between 1.025 - 1.027 for half a dozen tests in a row but just makes me uneasy that I'm getting different readings. But always starting high and gets gradually lower ?

What refractometer?

Sounds like it may be a temperature equilibration issue. I’d suggest taking a water sample, let it sit next to the refractometer for a while (an hour maybe) then try it and see if the same thing happens.
 
Agree with using a calibration solution....but that doesn't address the drifting numbers.

How old is the refractometer? And does the little swinging cover plate swing freely?

Might wish to check out this post:

Lubricate Your Refractometer
 
What refractometer?

Sounds like it may be a temperature equilibration issue. I’d suggest taking a water sample, let it sit next to the refractometer for a while (an hour maybe) then try it and see if the same thing happens.
Quite an old one judging by the box ! Will give that a go thanks randy I suppose temperature equilibration could make sense as the more tests I do i suppose the closer the temperature matches and generally more settled the results become. Maybe time to grab a new one to be sure regardless
 
Agree with using a calibration solution....but that doesn't address the drifting numbers.

How old is the refractometer? And does the little swinging cover plate swing freely?

Might wish to check out this post:

Lubricate Your Refractometer
Probably older than me by the looks of it that is swinging fine so that's one more thing to rule out but thanks for the insight
 
I really like the fauna marin hydrometer.
 
I really like the fauna marin hydrometer.
Looks good , May look in to that as I run my tank between 77 -78 so no need for the conversion chart
 