The mistake I keep making is pressing the button on the KH unit until the timer appears. It just turns itself off as there is no timer
These are the tips about Hanna checking that I've compiled from this forum (so far..)
***
For consistent results, always use the same vial and cap for every test.
Make sure the inside of the cuvette is free of stains and contaminants.
Rinse out the cuvette with tank water twice.
Use a 10 ml syringe to fill the cuvette with water (the fill lines aren't accurate)
(The bottom of the water curve should touch the fill line on cuvette)
Before placing the cuvette in the tester, dry and clean the cuvette thoroughly with a microfibre cloth before so it is crystal clear (hold it up to the light.)
Always place the cuvette into the tester with the same orientation (e.g., 10 ml mark to the front or hinge of the tester.)
Never place a reagent pipette contaminated with seawater in the reagent bottle. Never allow the pipette to touch the seawater sample.
Always replace the lid on the reagent bottle immediately after drawing the reagent.
Always open reagent pouches correctly (instructions are on the pouch.)
Shape the pouch into a funnel and tap as you pour into cuvette.
After shaking, tap and gently swirl the cuvette to dissipate all bubbles before testing.
Make sure there is no detritus in the cuvette before testing.
Do not leave the reagent in the cuvette too long as it will stain.
After the test, thoroughly rinse the cuvette and syringe pipette with RO/DI water.
Always store the cuvette filled to the brim with RO/DI water.
Once opened, the KH kit reagent will last approximately 2 months.
After 25 tests, discard the KH reagent bottle with any remaining reagent.