The answer is kinda, but not a good idea to use the hanna HR NO3 packets in the Cu checker. And there's a better solution available.
Here's calibration data for my tank water spiked with a few NO3 levels and run the hanna HR NO3 chemistry on it, and put it in either the hanna HR NO3 checker or the LR Copper checker.
| Total ppm NO3 stock | HR NO3 checker (ppm) | LR Cu checker (ppb) |
| 0.0 | 0.5 | 51 |
| 2.1 | 3.0 | 119 |
| 5.1 | 6.4 | 207 |
| 10.3 | 11.4 | 423 |
| 20.5 | 22.3 | 748 |
and while it looks good....
This is not a good wavelength for this test.
The LED wavelength for the Cu checkers is going to be out around 575nm which means you are measuring waaay on the edge ot the peak. Measuring that far out means that small differences (one LED's checker vs the next, one reagent lot vs another) will mess up the result a lot.
But, you should instead run a Red Sea NO3 test in the checker.
Red Sea NO3 color is much closer to the right wavelength for the Cu checker.
This is the spectrum of the Red Sea NO3.... measured by
@Rick Mathew here
as you can see, the 575nm hits very near the fat part of the peak and is a much better pair to use.
In fact, it looks like Rick may already have done this years ago.
This is regression data for measuring Red Sea NO3 with a hanna HR copper checker. (different model Cu checker, but I'll bet it works the same.)
Rick's approximate regression is about [NO3 ppm] = 6.0 x [Cu checker ppm].
@Red2143 you may give that a try.