That graphic above has some problems. To avoid confusion:
*Carbon blocks are radial flow, not axial.
*Membranes are not intended to remove bacteria. In fact, the membrane manufacturers specify that the membrane should be used only on microbiologically safe water.
*The DI stages should not be connected by a hex nipple. Rather, they should be connected by a loop of tubing to allow the user to mount a TDS meter probe between the DI housings.
*The graphic misuses the word "Stage." Each filter in the system is a Stage. I think they meant "probe."
*The pressure gauge in that pic is plumbed in after the third stage, not the second stage.
*The Triple probe TDS meter, TRM1, has the green face. The blue one shown is the DM1 which has two probes.
*On the system shown, you'd ideally want four TDS meter probes: 1) feedwater, 2) permeate, 3) after first DI, and 4) after second DI.
Russ