Different kit tables can be confusing to folks and some people do read them backwards.
The easiest way with any titration kit is to understand that the more liquid that must be added to get to the endpoint, the higher will be the resulting concentration from their table. So just make sure that you are reading it such that if you used only a few drops, the concentration value is very low, and if you had to dispense nearly a whole syringe of liquid (or more than one whole syringe), the value would be high.
Does that clarify it?
I don't have the Salifert table you are using, but if the syringe pushed all the way down to 0 gives a calcium value of 596 ppm from the table, then the table is showing the final syringe position (not the volume dispensed). So you take the number directly from the syringe for their table, not determining a value by subtraction.
The easiest way with any titration kit is to understand that the more liquid that must be added to get to the endpoint, the higher will be the resulting concentration from their table. So just make sure that you are reading it such that if you used only a few drops, the concentration value is very low, and if you had to dispense nearly a whole syringe of liquid (or more than one whole syringe), the value would be high.
Does that clarify it?
I don't have the Salifert table you are using, but if the syringe pushed all the way down to 0 gives a calcium value of 596 ppm from the table, then the table is showing the final syringe position (not the volume dispensed). So you take the number directly from the syringe for their table, not determining a value by subtraction.