Additional information.
I studied the Seneye performance in a 250 mL beaker with the pH probe about one inch away. The tank water was magnetically stirred. There was a litlle whirlpool in the beaker. I monitored the pH and temperature with a calibrated pH meter. What this all means is that I ensured the water visiting the Seneyed device was identical ti that which visited the pH probe. My experiments lasted less than 12 hours and I took readings 20-40 minutes apart. The temperature is typically low by tenths of a degree. For the three slides that I am studying, the pH is low by 0.1-0.4. The ammonia story needs to be covered separately, but I can say that it responds within 2-3 minutes and is fully equilibrated in 20 minutes.
Comparing a pH kit measurement to a Seneye pH might be tough The test kit variability might be as high as the difference between the two measurements. To remove as much variability as possible for a kit v Seneye measurement grst, put the Seneye in a pint of tank water. Stir it frequently over a one hour period. At the start, middle and end of the one hour period, measure the pH and temperature with the Seneye and your test kit and own temperature probe. This will really inform you about the difference.
As for the temperature difference between your probes and the Seneye, the fairer comparison is with the Seneye and your temperature probes next to each other. I bet you find the consistency with improve.
I don’t use the Seneye for remote monitoring. If I did, the alarm settings would not be set too tight to avoid nuisance alarms.
I hope this helps.