Here are the instructions for anyone else curious
Calibration Your monitor was factory calibrated to 342 ppm (NaCl). This level is suitable for most tap water/filtered water applications. However, you may need to re-calibrate based on your needs, as well as from time-to-time to ensure best results.
1. Purchase a calibration solution from your dealer that is correct for your needs.
2. Disconnect the T-Fitting from its hose. Do not remove the sensor from the T! Ensure the orientation of the sensor to the fitting is correct, as in the illustration.
3.Turn on the monitor and place the T-Fittings (with the sensor in it) into the calibration solution. You will get a reading.
4. If the reading on the monitor does not match the solution, adjust the reading up or down by gently turning the orange screw on the rear of the monitor clockwise or counter-clockwise.
5. The monitor should read approximately 10% below the calibration solution. The monitor is designed for flowing water and this discrepancy will compensate for that. (NOTE – if you are calibrating with a flowing solution, ignore this).
6. Once the reading is correct, turn the monitor off and remove from the solution.
Your monitor is now calibrated