Sudden Alk Rise

Users Who Are Viewing This Thread (Total: 0, Members: 0, Guests: 0)

shanedag

Active Member
View Badges
Joined
Oct 18, 2020
Messages
190
Reaction score
287
Location
Elmira
Rating - 0%
0   0   0
A bit stumped on this one. Using the Trident and have had really steady Alkalinity readings for couple months (right around 9.5-9.6. It's tested 3 times a day. Suddenly overnight, it jumped to 13.97 and has been slowly declining over the course of the 3 tests today? Could it really be that big of a jump over night?
 
It should test 4 times a day minimum. 2 things, check how low reagent is and also make sure the intake line isn’t slightly plugged. If it’s not filling the chamber enough it will give a stupid high reading. Happened to me twice since owning it. The line will not be plugged solid just some gunk in there.
 
Last edited:
It should test 4 times a day minimum. 2 things, check how low reagent is and also make sure the intake line isn’t slightly plugged. If it not filling the chamber enough it will give stupid high reading. Happened to me twice since owning it. The line will not be plugged solid just some gunk in there.
Man, thank you! I think you've nailed it. Fusion was reading that I had 34% left in the reagent A bottle, but I just checked and it was empty. Replacing and priming now. Hoping this clears it up.
 
That was definitely the issue. Levels back to normal on Trident reading. Thank you!
 