Warning about ICP procedures that can cause false readings: freezing

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
72,100
Reaction score
69,741
Location
Massachusetts, United States
Rating - 0%
0   0   0
I do not know what tube filling procedures are advised by all ICP companies, but a recent test by Christoph at Oceamo points out a potentially serious concern.

Freezing of seawater causes the formation of fresh water ice and hypersaline seawater. As the ice spreads to more and more of the total water, the saline liquid remaining can be very much concentrated relative to the original sample.

If all of the collected sample remains within the tube, there will be no change the in the element concentrations when the sample is thawed (at least not by the mechanism described below).

But if the tube is filled to near or at the top, having little air volume to compress, the pressure can rise quite substantially during freezing, and this rise can push out some of the hypersaline liquid. Since that liquid no longer remains in the tube when the sample is thawed, those ions are lost from the measurement, and the net concentration of the remaining fluid after thawing is reduced. Christoph showed that a tube filled to the top and frozen to -4 deg F showed a drop in calcium from about 420 ppm to about 300 ppm, magnesium from about 1250 ppm to 900 ppm, etc. he also showed very high ion concentrations in the exuded fluid

These effects are far too large to ignore.

Thus, if there is any chance your samples may freeze, be sure to NOT fill them all the way up. If your ICP company advises filling to the top, I'd suggest contacting them to see if it is OK to only half fill the tubes.

Here is Christoph's data:

 
Thanks for the information.
 
Hi Randy,

thanks for posting this in prominent position!

@all: Please do not fill the vials only halfway full - the labs will run into issues because of limited water volume. Having a air cusion of ~10% is sufficient. So fill a 15 ml vial only to the 13-14 ml and a 30 ml vial only to the ~27 ml mark. This is sufficient to prevent any leaking, which will occur when the vials are filled completely.

@RHF: Is there a way dropping you a PM?

Thanks and best regards,
Christoph
 
Could we just ship with one of those hot packs?
 
A hot pack lasts like 24-48 hours tops. Most of these tests take 7-10 days to get where they’re going. The hot pack won’t make the trip
 
Very good point, I do communicate this in the RM Facebook Group during the Winter season for the Reef Moonshiner's Testkits (Oceamo ICP-OES/MS).

Time to implement this detail into the Sampling Instructions.
Great work from Christoph and the Oceamo Team, thanks for sharing Randy!

1706130818193.png
 
I sent in an ATI ICP-OES test a couple weeks ago when it was cold. Salinity, Kh, Ca, and Mg were all suspiciously low compared to my Trident and hand tests. ATI commented that the RODI sample leaked so much they were unable to perform the RODI test. The samples must have frozen. I've always filled the tubes completely because I remember reading somewhere that the sample can interact with air.
 
IMO overkill and not necessary. Just dont fill your vials 100%, that is the only measure necessary.
Best regards,
Christoph
@Christoph I can not find this ICP kit in Canada. Do you have a timeline for availability?
 
Maybe a dumb question, but if the tube is upside down when it freezes (I doubt it can be kept vertical while shipping), won't some liquid be forced out when the air at the top expands/contracts? Or is the cap seal tight enough that nothing is forced out and the 10% of air just absorbs the pressure change?
 
Thanks for the info.
 
Maybe a dumb question, but if the tube is upside down when it freezes (I doubt it can be kept vertical while shipping), won't some liquid be forced out when the air at the top expands/contracts? Or is the cap seal tight enough that nothing is forced out and the 10% of air just absorbs the pressure change?
The tubes can withstand quite some pressure. The air in the tube just allows the liquid to expand without leaking, regardless of orientation of the tubes.

All the best, Christoph
 

IF YOU HAD TO TAKE A REEFING EXAM, WOULD YOU PASS?

  • Yes!

    Votes: 32 45.7%
  • Not yet, but I have one that I want to buy in mind!

    Votes: 9 12.9%
  • No.

    Votes: 26 37.1%
  • Other (please explain).

    Votes: 3 4.3%
Back
Top