I have actually stated numerous times that you are misstating and/or not understanding what I (and others) are trying to explain to you.
Several people here have attempted numerous times to help you understand what +/- .1 accuracy means in context to what your instrument displays and how that reading compares to the actual pH.
Try this:
If we take TWO +/- .1 pH probes that are guaranteed to be +/- .1 ph Unit of the actual fluid.
You measure a sample with YOUR probe and it reads 8.1
the actual value could be 8.0, 8.1 or 8.2
The range from 8.0 to 8.2 is .2
You don't know what value is correct, it could be anywhere between 8.0 and 8.2
I come to your house and measure the same sample with my +/- .1 probe.
my probe reads 8.2
the actual value could be: 8.1, 8.2 or 8.3
The range from 8.1 to 8.3 is .2
I don't know what value is correct, it could be anywhere between 8.1 and 8.4
Both probes are reading within their specified accuracy.
IMPORTANT:
You can't see my probe reading and I can't see your probe reading. But both probes are accurate to +/- .1
Which reading is correct and which probe is correct? We know have values between 8.0 and 8.4! I am sure that value is between 8.1 and 8.3 but you are sure the value is between 8.0 and 8.2
Are you beginning to see the problem?
Now if you and I compare values, we can see that the (if both probes are accurate) then the real pH is likely 8.1 or 8.2
Great, if we already know both probes are accurate. But if we don't then we are back to where we started.
That is - if your probe IS accurate, it is telling you within .2, not .1 what the pH is.