No sample water Trident error

  • Thread starter Thread starter Waters
  • Start date Start date
  • Tagged users None
Yes the clear silicon tubing is different from the opaque tubing. I recently had an issue with mine on the small bit of tubing between the metal sample needle in bottle A and the 90deg fitting. Alk measurements were all over the place. I took the cover off and watched what was happening and there were air bubbles in the line as it was adding A to the vial.. sometimes... I bought some 1.5mm ID silicon tube and replaced that small bit of tubing inside the bottle and no more air bubbles. I suspect the tubing that comes with the unit is getting old and the seal around those little barb fittings is not as good as it used to be.
 
This came out of my Trident yesterday for a no sample error
 

Attachments

  • CAA53524-D2F2-4B51-B298-6229EFD6E9CF.jpeg
    CAA53524-D2F2-4B51-B298-6229EFD6E9CF.jpeg
    156.6 KB · Views: 47
I recently had this error as well. Ended up being a clog in the manifold area. I used a titration syring from a salifert mag test kit to push RO water through the lines.

 
Massive props to Apex for the engineering of this thing.

The limits of my mechanical aptitude and fine motor skills are tested just connecting the USB cable.

Even I was able to disassemble this thing, fox the clog, and put it back together. I can’t believe it.
 