I doubt the milwaukee fluids are bad if only a couple month old and haven't been opened. If you are saving them after use then well.. they could be who knows what..
I messed my calibration up once doing a manual calibration through apex fusion and not letting the two readings settle properly before hitting next. I have also heard it is better to do the manual calibration vs the automatic mode. I also double check after going through calibration by retesting the probe's readings in both the 7 and 10 fluid. I set the apex to do a 1 minute refresh on readings and I leave the probe in until it gives me an updated reading.. AND I don't know if it helps but I dip the probe in a cup of rodi water to rinse before switching from between tank/ph7/ph10, followed with shaking off any drips.
i do exactly what you do. rinse probe w/ rodi, place in 7 ph solution, wait till apex says next step but i wait an additional minute. then rinse probe in rodi again, place in 10 ph solution and wait a bit before finishing. after calibration i place the probe back into ph 7 solution to make sure it gives me something around 7 ph and then back into 10 ph solution to get a 10 reading. i just recalibrated w/ the auto calibration function again w/ the Milwaukee solution and again i'm getting same PH readings in my tank. its sitting at 8.45 before my lights even come on.