I did than many times. Now I always wait at least 24 hours. The suspended salt may have thrown off your readings.Until the water turned to almost clear
Follow along with the video below to see how to install our site as a web app on your home screen.
Note: This feature may not be available in some browsers.
I did than many times. Now I always wait at least 24 hours. The suspended salt may have thrown off your readings.Until the water turned to almost clear
Sounds good I will do that thank you!Calibrate and recheck after 12-24 hrs. First calibrate with RO water at ZERO then solution at 35 ppt then try again with measurement
I will definitely check this out now that I think about I had my windows in the house open and the temp dropped to 58 degrees overnight so that probably my issue I will now more later this afternoon thank you!This sounds like a problem I have had. Consider whether the problem might be your refractometer—specifically, the heating and cooling of the internal optics as you handle it between measurements.
I use a terrible, cheap refractometer, but make it work. It supposedly has automatic temperature control, but the readings bounce all over the place even if I just warm it in my hand for 20 or 30 seconds.
To get around this, I check it with 35ppt calibration fluid every time I use it. Then I check my sample against whatever reading it gave for the calibration solution. If they’re the same, great. Otherwise, I know to adjust up or down. I try not to handle the refractometer too much (body heat), and I don’t bother actually calibrating the thing.
I run my tank at 35ppt, so I just want to know whether my water is as refractive as the calibration solution. Even a cheap, uncalibrated refractometer can tell you that.
Edit: try watching the reading as you hold the refractometer for a minute or two. See if the measurement drops (heating should cause it to read lower and lower). If you see that happening, I bet that’s the issue.