During all this discussion I had a postulate and that was that you have the right numbers from your Hanna. I have to change this now. You have the exactly same model as I have and I have been in exactly the same situation as you with one exception. My tank was just on the brink to crash when a Triton ICP test shows that my PO4 was 0.02 ppm - not 0.1 as my HI774 shows. Nowadays - I use my HI-774 in order to see if my PO4 level is constant or not and accept the figures it shown as long as my corals going fine. At least my example of the HI - 774 is rather unstable with it reading. The accuracy that should be ± 0.02 mg/L often show ±0.04 mg/L when I do several test after each other. Nowadays - I use 2 vial with test water. 1 that it is the zero and one that it is the sample. Before adding reagent to the sample tube - I control that both shows zero in calibration. I take the zero samples and do the process till C2 shows up. Switch the vials and do a short press on the bottom. With a short press it start to analyse direct - without the 3 minutes colouring up time. If it show zero - I start the process and add reagent to the sample vial. If not showing zero - I clean the vials again

I´m following the process with 2 minutes shaking - after that I zero the meter with help of the zero vial - when C2 shows up I switch the vials and do a long press - starting the 3 minutes colouring time. When I get an answer - I write it down. Now - a new zeroing with help of the zero sample. At C2 change the vials again and do a short press - starting to analyse directly. I repeat this till I have 10 readings - take the average.
I have notice that at least my HI 774 is very sensitive for how I shake the sample. If I shake it very carefully - I get a more steady reading and lower. at my job - I use HI 713 instead and I feel like that meter is more stable.
What I´m trying to say is that try to validate your results with another instrument or method - and look more at your corals instead for your meter.
Here is some threads where this is discussed
I bought an ATI ICP test for the first time. Does anyone know if outside temperatures when mailing them off affect the water samples?
www.reef2reef.com
Can someone please give me some good references on what I can expect to see with different phosphate levels? What happens to SPS in high phosphate levels? what happens to them in low phosphate levels? What happens to phosphate throughout the day? does it swing in some sort of sine wave? Does it...
www.reef2reef.com
SAMPLE STORAGE AND ITS IMPACT ON PHOSPHATE MEASUREMENT In a discussion on R2R @BigJohnny posed a question about why his Hanna ULR Phosphorous test results were significantly higher (2-3 times) than his ICP test results…. Here is a link to the discussion...
www.reef2reef.com
Sincerely Lasse