I am getting inconsistent readings. I checked my PO4 four times and got 0.135, 0.202, 0.0, and 0.0. I am usually 0.04 or less. I have not changed anything so I wonder what is going on?
Follow along with the video below to see how to install our site as a web app on your home screen.
Note: This feature may not be available in some browsers.
I am getting inconsistent readings. I checked my PO4 four times and got 0.135, 0.202, 0.0, and 0.0. I am usually 0.04 or less. I have not changed anything so I wonder what is going on?
I will try that. I clean my vial after use and even tried the second vial. I also place the vials in the test in the same position so the tester reads from the same loacation.If you have a second Hanna test vial, fill it with RODI or tank water. It will serve as the blank, or the vial you put into the Checker when “C1” flashes. Prepare a test sample with the usual mixing and hold times. The idea is to to retest the same sample over and over again, using the blank for “C1” and the same test sample for “C2”. If the Checker is the problem, instead of getting very similar readings each time, they might bounce around.
If the Checker behaves consistently, then the variation is coming from bad reagents, possible, not likely or a contaminated or dirty vial, which is more likely.
I have always used the same vial to zero the meter and test. I get the regents set to go and shake the vial until there is no residue. Has always worked until today.I had the same issue. Was advised to use the same vial to zero the meter and to do the actual test. Add C1, get it out quick, add reagent, shake, polish, put it back in for add C2. Mine is the LR, not the ULR. I'm no expert, but that was the overwhelming consensus when I tried the #askBRSTV fb page. I think it was even a hanna rep. Point being, it's a PAIN cuz mine times out after 3 mins and getting the regents out of that little envelope and shaking for two minutes, plus buffing the vial is hard. Even so, my results have been a lot more consistent since using that approach. Still not what I'd like, but I'm not questioning the trends anyhow.
I don't know whyI have always used the same vial to zero the meter and test. I get the regents set to go and shake the vial until there is no residue. Has always worked until today.
I would be concerned that the dishwasher leaves a lasting PO4 residue given the wash and drying cycle. I don't think the 2 minute shake time is critical, the reagent just need to be totally dissolved. I have been doing this weekly, same procedure, for over a year and this is the first problem. The previous tests have been as expected.I use two cuvetts, that I clean in the dishwasher and rinse with distilled water before I test. One for the test sample filled with tank water and the blank filled with tank water to solve the time out issue. I use a magnetic stirrer to stir the test sample for two minutes. I wipe down the blank with a clean napkin and place the blank in the Hanna unit while the sample is stirring. Once the same has it 2 minute stir, then I place the sample in the unit and press for the 3 minute timer on the unit to start. I wait and get my results.
Make sure the cuvetts are clean and the same is shaken/stirred for the 2 minutes.
If you have a second Hanna test vial, fill it with RODI or tank water. It will serve as the blank, or the vial you put into the Checker when “C1” flashes. Prepare a test sample with the usual mixing and hold times. The idea is to to retest the same sample over and over again, using the blank for “C1” and the same test sample for “C2”. If the Checker is the problem, instead of getting very similar readings each time, they might bounce around.
If the Checker behaves consistently, then the variation is coming from bad reagents, possible, not likely or a contaminated or dirty vial, which is more likely.
Hey everyone. You must have the Checker analyze the sample exactly 5 minutes after the powdered reagent is added and mixing begins. Once that 5 minute mark passes, the color fades. So, for about 5 minutes the color develops and after 5 minutes it fades (I.e. lambda-max is at 5 minutes).
For repeatability and reproducibility, mix for exactly 2 minutes (time it with phone stop watch) and start the checker’s 3 minute on-board timer right at the 2 minute mark.
The above was obtained by @MnFish1 directly from Hanna.
How to u get magnetic stirrer out of vial??I use two cuvetts, that I clean in the dishwasher and rinse with distilled water before I test. One for the test sample filled with tank water and the blank filled with tank water to solve the time out issue. I use a magnetic stirrer to stir the test sample for two minutes. I wipe down the blank with a clean napkin and place the blank in the Hanna unit while the sample is stirring. Once the same has it 2 minute stir, then I place the sample in the unit and press for the 3 minute timer on the unit to start. I wait and get my results.
Make sure the cuvetts are clean and the same is shaken/stirred for the 2 minutes.
How to u get magnetic stirrer out of vial??
I'm thinking of using one.
leaves a lasting PO4 residue
So I just bought a replacement ULR, does that mean I should now shake for 2 and sit for 5 mins?FYI, speaking with Kevin Costa at RAP, Hanna is re timing the new ULR and reagent for 7 minutes instead of 5.
Hence the rinse with distilled water. I have not had an issue, Glass will not saturate.
So I just bought a replacement ULR, does that mean I should now shake for 2 and sit for 5 mins?